Choosing the right sensor chip for a low‑affinity small‑molecule ligand
- zeduzeni1992
- Topic Author
- Offline
- New Member
-
Less
More
- Posts: 3
- Thank you received: 0
1 day 15 hours ago - 1 day 6 hours ago #1
by zeduzeni1992
Choosing the right sensor chip for a low‑affinity small‑molecule ligand was created by zeduzeni1992
I’m trying to measure the interaction of a low-molecular-weight drug-like molecule with a protein that has a relatively high micromolar K_D. Which sensor chip (e.g., CM5, SA, NTA) would give the best signal-to-noise, and what regeneration strategy would you recommend to avoid losing the immobilised protein? Any practical tips from similar experiments are welcome.
Last edit: 1 day 6 hours ago by SPRAdmin04.
Please Log in or Create an account to join the conversation.
- SPRAdmin04
-
- Offline
- Administrator
-
Less
More
- Posts: 3
- Thank you received: 0
1 day 6 hours ago #2
by SPRAdmin04
Replied by SPRAdmin04 on topic Choosing the right sensor chip for a low‑affinity small‑molecule ligand
Main issue is the size difference between the ligand and analyte. Aiming for an analyte response between 20 and 50 RU you can calculate the amount of ligand to be immobilised. Using a high performance machine you even can aim to a lower Rmax. If the small compound dissociates within 10 - 15 minutes i would do no regeneration step but just wait and saving your ligand.
Please Log in or Create an account to join the conversation.