Abnormal +4000 RU Baseline After EDC/NHS Activation with No Ligand Immobilizatio
- charlie16
- Topic Author
- Offline
- New Member
-
Less
More
- Posts: 1
- Thank you received: 0
4 days 5 hours ago #1
by charlie16
Abnormal +4000 RU Baseline After EDC/NHS Activation with No Ligand Immobilizatio was created by charlie16
Here is my current situation:After EDC/NHS activation, the RU does not return close to baseline. Instead, after washing it stabilizes around baseline +4000 RU, whereas tutorials show it should typically remain around baseline +200–300 RU. During ligand injection, I also do not observe the typical gradual RU increase — the signal remains almost flat. However, my preconcentration test looked normal.Reagent details:
- EDC: 750 mg dissolved in 10 mL deionized water (self-prepared)
- NHS: 115 mg dissolved in 10 mL deionized water (self-prepared)
- Ligand diluted in sodium acetate buffer, pH 5.0
- Ethanolamine prepared from pure reagent and adjusted to pH 8.5 with HCl
- After EDC/NHS activation, what RU level is normally expected after washing?
- Is it acceptable to prepare EDC and NHS stocks manually in DI water like this, or could that be causing abnormal activation behavior?
- Has anyone seen a similar baseline behavior after EDC/NHS activation? Any suggestions on likely causes or troubleshooting steps would be greatly appreciated.
- Could this abnormal high baseline (+4000 RU) be caused by hydrolysis of EDC and formation of side products (e.g., urea derivatives) that accumulate or adsorb on the surface?[img]IMG_1303.jpeg IMG_1301.jpeg IMG_1302.jpeg[/img]
Please Log in or Create an account to join the conversation.
- Arnoud
-
- Offline
- Moderator
-
Less
More
- Posts: 46
- Thank you received: 6
3 days 8 hours ago #2
by Arnoud
Replied by Arnoud on topic Abnormal +4000 RU Baseline After EDC/NHS Activation with No Ligand Immobilizatio
Short answers:
1): normally between 125 - 250 RU on a dextran based surface (CM5)
2): Yes, I dissolve a full bottle and then aliquot in single activations (150 uL) and freeze at -20 celcius for up to 2 years.
3): Not the high baseline after immobilization.
4): I would say no, and they will be washed away by the deactivation.
Unfortunaly the images where not uploaded. Can you try again?
regards
Arnoud
1): normally between 125 - 250 RU on a dextran based surface (CM5)
2): Yes, I dissolve a full bottle and then aliquot in single activations (150 uL) and freeze at -20 celcius for up to 2 years.
3): Not the high baseline after immobilization.
4): I would say no, and they will be washed away by the deactivation.
Unfortunaly the images where not uploaded. Can you try again?
regards
Arnoud