SPR-Pages

pagelogo
back
problems coupling anti-GST - SPRpages
back
These are the posts of the old forum. It was not possible to transfer the user data, so they are missing in most of the posts. For new questions, go to the general discussions.

problems coupling anti-GST

  • OldForum
  • Topic Author
  • Visitor
  • Visitor
14 years 1 week ago #1 by OldForum
problems coupling anti-GST was created by OldForum
Hi, I'm trying to use the GE healthcare GST capture kit with a CM5 chip

The kit has anti-GST and immobilisation buffer and I'm trying to NHS-EDC couple the anti-GST to the chip
The kit estimated I should be getting 7500 RU bound. This allows me to then couple about 500 RU of GST-fusion to this to then bind to my analyte

Problem is, I can see a really big peak when I flow the anti-GST over - around +30000 RU
But most of this doesn't leaving me with about 500RU on the chip.

Not sure what the problem is, I've tried lower flow rate and high [] of the anti-GST


Machine is a biacore 3000

Please Log in or Create an account to join the conversation.

  • OldForum
  • Topic Author
  • Visitor
  • Visitor
14 years 1 week ago #2 by OldForum
Replied by OldForum on topic problems coupling anti-GST
A peak of >30.000 RU indicates that there is air injected instead of liquid, or somehow there is a high salt/glycerol concentration in the injected liquid.
I would say: check the injection system.

To check this, try injecting flowbuffer: it should give a minor baseline displacement of max 5-10 RU.

Arnoud

Please Log in or Create an account to join the conversation.

Powered by Kunena Forum