Dear all,

Can anybody give me his opinion in the reason why my sensorgram shows a small increase after injection which is followed by small decrease and completion of association and dissociation phases afterward (as shown in the attached figure specially
with the lower conc)?

Attachments:

Please Log in or Create an account to join the conversation.

Sensorgram increase at the start 05 Jan 2016 10:21 #2

Hi Ahmed,

The spike come from the fact that the channels are in series. When subtracting a reference, both channels are a little out of phase.
In the Biacore you can use the direct referencing during the experiment which will give less spikes.
Or you can adjust the injection time point t=0 for each curve.

When using a program like Scrubber, you kan artificially remove the spikes.

Large spikes occur mainly when the bulk effect is large (>25 RU) and the compensation by the reference channel is not optimal.

Kind regards

Arnoud