TOPIC: Temperature Problem

Temperature Problem 27 Feb 2015 11:40 #1

Hallo everyone,

i am working with a BIAlite ugrade system. It worked fine for 20+ Years. However since wednesday the system displays a temperature failure.
The temperature is not longer constant. It varies from 24 to 27°C. The room temperature is constand at 22°C. I looked in the handbook, but there is no hint how to solve a problem like this.
Does anybody have an Idea or a suggestion how to repair the maschine?


Temperature Problem 01 Mar 2015 10:59 #2

For me looks like an hardware problem.
Temperature sensor is broken or a malfunction in the cooling/heating element.
Contact the service to ask if you can do any diagnostics before they come.
Probably you have to replace some parts ...

The following user(s) said Thank You: DAzureus

Temperature Problem 02 Mar 2015 11:01 #3

Thanks for your answer.
For me it looks like the sensor. It changes from 24 to 27 °C in one second.
I disconnected the sensor, used contact spray and connected it again. Now i have to wait a little bit to see an effect
Service for this old system (25 Years) is no longer available :(