TOPIC: Reference surface

Reference surface 19 Mar 2012 01:00 #1

  • OldForum
  • OldForum's Avatar
    Topic Author
  • Offline
  • Posts: 414
Hi everybody,

I have a doubt. I have biotynilated antibody that I will capture over a streptavidin chip. I dont have any other biotnylated antibody that I could use for the reference spot. What would you recommend in this case? a bare Streptavidin spot? Activate and deactivate the chip? Covalenty immobilize an antibody on the surface?

Regards,
joseantonio

Please Log in or Create an account to join the conversation.

Reference surface 19 Mar 2012 01:00 #2

  • OldForum
  • OldForum's Avatar
    Topic Author
  • Offline
  • Posts: 414
Hi,

I am using now a streptavidin chip and use as reference just the native surface with streptavidin on it. There is no background binding so this is ok.

Activating / deactivating is not usefull because you don't do that with the other surface. Actually it is immobilized with streptavidin. Probably you will introduce some modifiaction you do not want with an extra activation.
It is maybe better to saturate the reference surface with biotin.

Arnoud

Please Log in or Create an account to join the conversation.

Reference surface 19 Mar 2012 01:00 #3

  • OldForum
  • OldForum's Avatar
    Topic Author
  • Offline
  • Posts: 414
I just saw what you mentioned, Arnout. On an activated-deactivated spot I get more background than in SAv-coated spot. I have D-biotin (Sigma) for blocking the SAv, does anyone know what concentration should I use?

Regards,
joseantonio

Please Log in or Create an account to join the conversation.

Reference surface 19 Mar 2012 01:00 #4

  • OldForum
  • OldForum's Avatar
    Topic Author
  • Offline
  • Posts: 414
Start using 10 ug/ml and sature the surface. The response will be rather low (D-biotin is very smaal). Wait until it level out and wash thorougly befor switching to the other channels.

regards
Arnoud

Please Log in or Create an account to join the conversation.

Reference surface 19 Mar 2012 01:00 #5

  • OldForum
  • OldForum's Avatar
    Topic Author
  • Offline
  • Posts: 414
Did it worked? If not, putting BSA (0,1 mg/ml) into the running buffer might help.