TOPIC: non specific binding issues

non specific binding issues 05 Jun 2011 02:00 #1

  • OldForum
  • OldForum's Avatar
    Topic Author
  • Offline
  • Posts: 414
anyone, we think some non-specific binding is occuring and what to try out adding more P20 surfactant to the existing HBS-P buffer but not sure how much to add. any suggestions? currently the HBS-P has 0.5% v/v of P20 in it. Also would adding BSA to samples help? Any suggestions would be most helpful.....Thanks.
mtaraca

Please Log in or Create an account to join the conversation.

non specific binding issues 05 Jun 2011 02:00 #2

  • OldForum
  • OldForum's Avatar
    Topic Author
  • Offline
  • Posts: 414
0.5% is alrealdy quite a lot. You could try adding BSA to the running buffer (say 0.5 mg/ml for a start) and some extra salt (0.5M) or dextran. Working at low concentrations can also help.
joseantonio